Hi Chris,
Would it be possible to modify the 9-point calibration to 3-point calibration? 3 points should still be enough to get a reasonable fit to start with. It would provide a smoother transition from monkey barely fixating at one point to fixating at all 9 points.
Thanks,
ā
Sujay
Hi Sujay,
The standard eye calibrator fits a second-order function that has six parameters, so you need at least six samples to perform a fit. However, the linear eye calibrator fits a first-order function with three parameters, so three samples would be sufficient to get a fit with it.
Cheers,
Chris